No current with phosphate buffer

Discussions about CE, MEKC and related topics

3 posts Page 1 of 1
Hi all,
I want to separate some basic componds. I prepared 100 mM phosphate buffer (pH 2.0) and I used 50 um capillary (48.5 cm). Everything was OK, I got peaks. Now I have to use another CE instrumetation (I usually use it for anion analysis and it works perfectly).
But suddenly there is no current at all at +20 kV or the currrent is too low (4-5 uA). Moreover, if I observed current it desappeared in 5-10 min.
I tried to remade the buffer, to exchange to capillary (both 50 and 75 um), to flush longer, to vary the applied voltage. No stable current, no peaks.

Could you please recommend something?

Thanks
If you say there is no problem with the instrument, i can only think at dirty electrodes or bubbles/particles in the buffer. Have you degassed and filtered the buffer?
Man this is an old topic. :) I used to have a lot of problems keeping the capilliary on the cartridge without it touching and causing a leakage current. A trick I would do is to tape the entire capilliary with a piece of scotch tape to insulate it. After this you can wind it anyway you want to.
3 posts Page 1 of 1

Who is online

In total there are 0 users online :: 0 registered, 0 hidden and 0 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 0 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry