Analysis of Hydroxylamine: HPLC or GC?
Posted: Mon Sep 26, 2005 10:50 pm
by Apple_insci
I am about to develop a method for analyzing hydroxylamine. Has anyone ever analyzed it? Shoud it be done by HPLC or GC? Any suggestion or idea is appreciated.
Thanks!
Apple
Posted: Tue Sep 27, 2005 12:47 am
by Uwe Neue
For such a polar and basic compound, HILIC on a silica column is probably best, with detection at 210 nm and acetonitrile as the organic modifier.
Posted: Tue Sep 27, 2005 5:25 am
by Apple_insci
Thank you for your suggestion! But I am not familar with HILIC. Can our Waters Alliance or HP 1100 still be used? Which column do you recommend? Very appreciate your input!
Apple
hydroxylamane
Posted: Tue Sep 27, 2005 5:56 am
by philippem
Have you searched the web?
try
www.google.com or the application notes given by several suppliers for chromatographic columns. Can be found on their website
Good luck
philippem
Re: Analysis of Hydroxylamine: HPLC or GC?
Posted: Sat Oct 01, 2005 11:21 am
by S.M.PANDYA
Thanks!
Apple
Dear Friend
According to me for Polar compound You may try with Atlantis dC18 HPLC column with Acetonitrile : Methanol 90 :10 mobile phase .
Posted: Sat Oct 01, 2005 6:24 pm
by Uwe Neue
HILIC is the inverse of RP. You use a very polar stationary phase, such as the Atlantis HILC silica column, with a mobile phase with a large portion of organic (acetonitrile) and a small portion of water. You may actually try first to run a gradient from 95% acetonitrile / 5% water to around 50% acetonitrile / 50% water. Your analyte should elute somewhere in between these two compositions. With acetonitrile and 210 nm detection, you should see your compound.
Posted: Mon Oct 03, 2005 11:51 pm
by Apple_insci
Thank you all for your information. This is why I love this forum. I get fast education. I am purchasing Atlantis dC18 and HILIC column and will give them try!
Best Regards,
Apple
Posted: Tue Oct 04, 2005 4:37 am
by james little
what about reacting with an aldehyde to get an imine. Could do with aromatic one and get a chromophore and do with UV detector? plus get it to retain well..
Posted: Fri Oct 07, 2005 5:37 pm
by Einar Ponten
The beauty of this forum is that you get alternatives as well.
ZIC®-HILIC is a column dedicated for your problem.
http://www.sequant.com/products/products.php
It is available from SeQuant through The Nest Group,
www.nestgrp.com - the experts on HILIC!