Page 1 of 1
qualification hplc agilent
Posted: Thu Mar 05, 2015 11:02 am
by Besma2
hi everyone :oops:
who make a qualification of Agilent HPLC 1200/ 1100
who can help me to handle OQ / PV programe ChemStation
I have the method and sequence but I CAN'T start the test and determine the limits
it's urgent please , where I can find a link or document
thank you
Re: qualification hplc agilent
Posted: Thu Apr 02, 2015 5:48 am
by Alfred88
Hi.
I paraphrase your question:
Who can perform the qualification IQ/OQ for Agilent system, to meet the manufacturer's specifications, using Chemstation?
Of course, Agilent service engineers can help. Give them a call.
The qualification task will be more difficult than you think. (I have been there). Please consider to get help.
Posted: Thu Apr 02, 2015 8:14 am
by kubowicz.tomasz
Hello
If you have Chemstation with "Verification" tab you can do it on your own. All you need is:
-certified flowmeter (or volumentric flask and watch)
-certified thermometer
-standards for UV/RI/FLD detector
Regards
Tomasz Kubowicz
Posted: Mon Apr 15, 2024 4:12 pm
by Emefede
Hello everyone,
I have an Agilent 1100 system with Chemstation B.04.03 and I want to do an OQ/PV but the methods and sequences are not included, where could I get them?