The install on the Windows 95 computer uses version A05.02 of the ChemStation software. The modules in the 1100 stack are:
G1315A (connected to the PC)
G1316A
G1313A
G1311A
We have an 82350B PCI card that we can use for this, along with version A10.01 of the ChemStation software. I have installed this, along with the L02.01 Agilent IO libraries (the one that came on the CD for A10.01) on the Windows XP system, with no issues at this point. I ran IO config and set it as follows (as per the install doc):
SICL = hp82341
VISA = GPIBO
Logical Unit = 7
Bus Address= 30
System Controller = Checked
During the install of the ChemStation, I entered our license keys for the LC ChemStation and Spectral Evaluation. I ran the Configuration Editor, selected the hp82341 card, and added the 1100 System Access under Modular 3D LC System with address of 26, which is the same as what was used on the old system. As far as I can tell, this is all that's needed to establish communication with the 1100.
When I start the Instrument 1 Online app the module configuration screen appears and after a several second delay, comes up with an error:
"Module's &%@&@-CU's could not be opened. Check firmware version"
This error repeats itself if I click ok to it. The &%@&@ symbols change with each error message, but otherwise it's the same.
I've tried v15 of the IO libraries as well with the same result. In this version, in the Connection Expert screen the instrument attached to the PCI card shows up as GPIB0::26::INSTR, with a warning on it about the IDN string not being valid. If I run the Interactive IO app and send it the "*IDN?" command, it just returns "«x00»«x03»?". I don't know if this is how the 1100 is supposed to show up in here or if this is part of the main problem.
My best guess here is the firmware on the 1100 is not a sufficient level to run with either the 82350B PCI card, or the A10.01 ChemStation software, but I haven't found anything that says this setup requires a specific firmware level. I'm rather hesitant to try and update the firmware on this unit given how old it is and expensive it would be to replace if anything went wrong, so I'm hoping there is just some configuration problem somewhere that I've set wrong.
I don't think I can check the current firmware on the unit right now without an RS232 cable attached (which I would need to obtain) so I wanted to post in here in the hopes that somebody that knows this system better can tell me what's going on. I'm grateful for any help you can provide, please let me know if you have any questions and I'll try my best to answer them.
Thanks.