I compared EI and CI directly by switching only the source ion volume and turning on the the CI gas (iso-butane) between consecutive injections of the same test standard. I cannot recall the difference off hand, and I am not at the lab now, but it was enough to need extra concentration of the sample and larger injection volumes to see useful peaks in CI.I'm not familiar with your instrument, but the sensitivity trouble spots in GC-MS are generally, in order in which encountered by the analyte molecules,
1. dirty injection liner
2. dirty first inch of column (or not a clean cut across the tubing)
3. dirty last inch of column (ditto)
4. cold transfer line
5. sagging filament (or heavily oxidized)
6. low CI gas pressure
7. dirty source
8. dirty focussing lenses
9. dirty rods at entrance to quads.
10. dying EM
5) and 10) have already been mentioned. Did I forget the septum ??