Carryover?

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Currently working on a method that involves testing liquid formulations for trace level impurities.
Toluene is used in my bracketing standards as a reference for calculating concentration of impurities.
Everything was fine up to a point and then all of a sudden I started to get the toluene peak in my samples. The strange thing is that the peak area of the toluene in the samples is the same as in the standards so clearly this is not a build up of carryover over time.
What could be causing this?
Toluene in the samples would be the simplest explanation, but quite a coincidence that it has the same consentration as the the standards.

Details of what you are doing, and how would help, i.e. temperatures, flows, sample and standard prep and so forth.

Peter
Peter Apps
2 posts Page 1 of 1

Who is online

In total there are 185 users online :: 1 registered, 0 hidden and 184 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot] and 184 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry