Has anybody encountered an issue with the split ratio wondering on some methods?
Yes.

Welcome to the forum.

Peter
Peter Apps
You dont have any theories as to the cause, it seems almost random, I'm the only operator so I know its not another human?
Hi chricol

If you stick around you will find that useful answers come more quickly if you put some details into your questions, so that other people have a clear idea what it is that you need to know.

The potential causes for fluctuating split ratios include poor gas mixing in the inlet (particularly relevant when doing headspace injections), inappropriate choice of inlet liner and packing, inconsistent injection volume, poor choice of solvent, instrument malfunction, particular combinations of solvent, inlet temperature and oven temperature ..... the list goes on.

So what was your question again ?

Peter
Peter Apps
Thank You Your right my enquiry is kinda sketchy, the issue can be best described as for a given method on EZ chrom using a varian 3800 GC the method can be saved with a 30 split ratio lets say and week later,or so, when running a sample the result would be wacky and upon investigation the split ratio will have change to perhaps 10. This happens on 2 different varian 3800 controlled by 2 different computers.
Does this:

"The potential causes for fluctuating split ratios include poor gas mixing in the inlet (particularly relevant when doing headspace injections), inappropriate choice of inlet liner and packing, inconsistent injection volume, poor choice of solvent, instrument malfunction, particular combinations of solvent, inlet temperature and oven temperature ..... the list goes on."

provide any clues as to what might be going wrong ?

And does this;
Thank You Your right my enquiry is kinda sketchy, the issue can be best described as for a given method on EZ chrom using a varian 3800 GC the method can be saved with a 30 split ratio lets say and week later,or so, when running a sample the result would be wacky and upon investigation the split ratio will have change to perhaps 10. This happens on 2 different varian 3800 controlled by 2 different computers.
provide any information relevant to any of those potential causes of trouble ?

"The more you don't tell us, the more we can't help you."

I have a specific question; how do you know that the split ratio has changed ? And I still need to know the details of the operating conditions.

Peter
Peter Apps
6 posts Page 1 of 1

Who is online

In total there are 191 users online :: 0 registered, 0 hidden and 191 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 191 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry