6550 QTOF problems with sensitivity

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hi,

I turned on 6550 iFunnel QTOF after a long shutdown state. Vaccum builup took a long time but now I am able to operate it. However, there is steep dip in sensitivity in positive mode. I used to observe 0.5 M counts average intensity, now it is in few thousands. When I tuned the machine in positive as well as negative mode, the issue did not get resolved.

Further, I prepared fresh tune mix, cleaned source and nebulizer.

Please provide suggestions to improve the response of the MS.

Thanks,
SAM
Firstly I would clean thoroughly the source, replacing key parts as the electrospray needle. If it does not do the job, I would clean the MS inside, as per maintenance manual. Bu I do know that microplate detectors and electromultipliers do not like to be exposed to air (or anything else than vacuum) so maybe that is the problem
2 posts Page 1 of 1

Who is online

In total there are 21 users online :: 2 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Google [Bot] and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry