New user here. Have an issue that I was hoping someone would help me with. I am running EPA 8260C VOAs on an Agilent 5975B MSD with an Agilent 6890N GC. Attached is an Archon autosampler with an OI Analytical 4560 Purge and Trap. About 2 weeks ago, I took the mass spec down and cleaned source and changed filaments. Normal run of the mill maintenance. After I pumped down and started running again, I noticed a very annoying and bothersome "hump" at the beginning of my analysis. It comes after the solvent peak (methyl alcohol) and right in the range where my 8260C gases come out (around 1.85 minutes to 2.3 minutes). It is playing havoc with my gases. Does anyone have any ideas as to how to get rid of it or any experiences with this? I have attempted the following things in this order: 1)changed the water management valve on the purge and trap. 2)changed septum and liner on GC injection port  3) Installed a new column. 4) Installed a new trap.  None of these things have solved the problem. I have tried a couple of different diagnostic steps with uncertain results. First, I direct injected a standard to see if it was coming from outside the GC or was a GCMS issue. 2) I raised the mass threshold to higher than 35 amu as the library was identifying it as hydroxylamine (MW=31). The direct injection helped, but just a little. Changing the mass threshold also helped, but not enough. Does anybody have any ideas as to what is causing this "hump"?  Thanks preemptively.
G. Ornelas