Peaks on hexane blank run

Posted: Wed Feb 18, 2015 10:06 pm
by BobCA
ran the GC FID after former operator left. Maintenance log not found.

Image

GC FID, SSL inlet

Hexane blank - made sure all blanks and washes were fresh, along with fresh vials.

What are the first steps I need to take? Maintenance or software troubleshooting?

4 blanks run, when blanks are overlayed, matching peaks/consistent.

Thanks, in advance.

Re: Peaks on hexane blank run

Posted: Wed Feb 18, 2015 11:31 pm
by James_Ball
Check to make sure that the data being recorded in the method is FID on the proper channel and not the test signal. I believe in some of the software you could set the active data channel to a test signal for diagnostic purposes and it could be that since the pattern is so perfect.

Re: Peaks on hexane blank run

Posted: Thu Feb 19, 2015 6:56 am
by dblux_
ran the GC FID after former operator left. Maintenance log not found.

GC FID, SSL inlet
Then why you post in GCMS subforum ?
Hexane blank - made sure all blanks and washes were fresh, along with fresh vials.

What are the first steps I need to take? Maintenance or software troubleshooting?

4 blanks run, when blanks are overlayed, matching peaks/consistent.

Thanks, in advance.
Make so called blank run ie. run without injection. Check if peaks present. If no than check your hexane.

Re: Peaks on hexane blank run

Posted: Thu Feb 19, 2015 9:34 am
by Peter Apps
That is a software-generated test chromatogram. You are on the wrong channel.

Peter

Re: Peaks on hexane blank run

Posted: Thu Feb 19, 2015 3:35 pm
by dblux_
That is a software-generated test chromatogram. You are on the wrong channel.
Peter
Definitely. This is a test plot generated by GC. The area of 2nd peak is 10 times lower than 1st peak and the area of the 3rd peak is 100 times lower than 1st peak.
No contamination can make such regular pattern. Unfortunatelly I replied from my office where all picture sharing servers are blocked.