column conditioning needed or it is another problem?
Posted: Thu Feb 05, 2015 6:01 pm
I have a trace ultra gC which has been unused fro more than one year. The TR-225 column was not taken away from the GC although the GC had been off during this long period of time, (meaning it had stayed connected in the GC).
First thing i have to do is column conditioning.
I know the following steps should be followed:
1. Turn on the carrier gas and check there is a good flow. This was tested by dipping the detector end into
some methanol and checking for bubbles.
2. Allow the column to purge for twenty minutes to expel any oxygen that has entered the column.
3. Heat the column from 50 °C to the maximum temperature (230 C) at 5 °C/minute and hold for 1 hour.
Now i have the following questions:
1. What should be the flow of N2 carrier gas and temperature on the inlet? (Should there be any temp on the inlet?)
2. Should i disconnect the column end from the FID detector?
3. Should the detector be 'Flame On' during this time and its gases flowing in it or should it be OFF. the manual says the following but it is confusing me: ''If using a non-MS detector monitor the baseline during this time. If the baseline is still gradually falling after 1 hour consider extending the hold until a stable baseline is achieved. If using longer columns (> 30 m) or with thicker stationary phases (> 0.5 µm) conditioning may take longer due to the increased amount of stationary phase present'' <--- DOES this mean that the column should be connected to the detector?
Many thanks in advance.
First thing i have to do is column conditioning.
I know the following steps should be followed:
1. Turn on the carrier gas and check there is a good flow. This was tested by dipping the detector end into
some methanol and checking for bubbles.
2. Allow the column to purge for twenty minutes to expel any oxygen that has entered the column.
3. Heat the column from 50 °C to the maximum temperature (230 C) at 5 °C/minute and hold for 1 hour.
Now i have the following questions:
1. What should be the flow of N2 carrier gas and temperature on the inlet? (Should there be any temp on the inlet?)
2. Should i disconnect the column end from the FID detector?
3. Should the detector be 'Flame On' during this time and its gases flowing in it or should it be OFF. the manual says the following but it is confusing me: ''If using a non-MS detector monitor the baseline during this time. If the baseline is still gradually falling after 1 hour consider extending the hold until a stable baseline is achieved. If using longer columns (> 30 m) or with thicker stationary phases (> 0.5 µm) conditioning may take longer due to the increased amount of stationary phase present'' <--- DOES this mean that the column should be connected to the detector?
Many thanks in advance.