I have prepared some Methanol and Water mobile phases with pH 10.7 phosphate buffer (3.9 g/L HNA2PO4*2H2O brought to pH with NaOH). I am running this on a new Kinetex Evo core-shell C-18 with guard column and VWR detection at 254nm... trying to figure out why I am seeing the baseline that I am.
My mobile phases are
A) 10% MeOH, 90% phosphate buffer
B) 90% MeOH, 10% phosphate buffer
The gradient is
time
0: 100%A
20: 100%B
30: 100%A
35: 100%A
I'm also seeing that the back pressure follows the gradient, as you would expect, but instead of being highest with the highest percent aqueous, it is maximized at about 50%A/50%B.
I'm hoping there is no precipitation when the phases mix, but I'm wondering if this is possibly happening--and my guard column is catching it and causing the increased back pressure.
I still can't explain why the dramatic change in baseline, though.

Any ideas? This also happened at 277nm.
Thanks in advance!