Dissolution samples in SLS media
Posted: Fri Sep 23, 2005 1:44 pm
Dear Members
I need to perform the HPLC analysis of samples of dissolution study of a lipophilic drug. The dissolution media need to contain 0.5% w/v of Sodium lauryl sulfate. I can at maximum dilute sample by equal volume of solvent in consideration of drug concentration in diluted samples. Thus reducing SLS to 0.25% w/v. Which probably is very high concentration. Injecting these samples change the peak shape badly specially downward (a sort of shoulder).
SLS possibly (in fact it does) accumulate on the column (C18). Distorted peaks are observed with the 'pure drug (without SLS) injected after 'samples containg SLS'.
Similar is the case with the hydrolytic degrdation product of drug.
I got to know 'cleaning method for SLS for column' in a post in this forum. However, SLS is practically making method not feasible for samples. I googled for similar problem and got this link.
http://dissolution.com/vbulletin/archiv ... t-258.html
I am putting a part of that in this forum as under.
I look forward for possible solution for such case.
I remember previously we developed a method containing Sodium octyl sulfonate as ion pairing agent and injected samples of 'same drug containg SLS' (in fact dissolution samples only). We had not observed any peak distortion in that case.
I look for the comments of the forum Members.
I need to perform the HPLC analysis of samples of dissolution study of a lipophilic drug. The dissolution media need to contain 0.5% w/v of Sodium lauryl sulfate. I can at maximum dilute sample by equal volume of solvent in consideration of drug concentration in diluted samples. Thus reducing SLS to 0.25% w/v. Which probably is very high concentration. Injecting these samples change the peak shape badly specially downward (a sort of shoulder).
SLS possibly (in fact it does) accumulate on the column (C18). Distorted peaks are observed with the 'pure drug (without SLS) injected after 'samples containg SLS'.
Similar is the case with the hydrolytic degrdation product of drug.
I got to know 'cleaning method for SLS for column' in a post in this forum. However, SLS is practically making method not feasible for samples. I googled for similar problem and got this link.
http://dissolution.com/vbulletin/archiv ... t-258.html
I am putting a part of that in this forum as under.
Considering the our case its not possible to keep the column aside as we need to perform very few studies compared with average column life time span."Use of surfactants in hplc is well known-called soap chromatography or dynamic ion exchange chromatography. The SLS will be adsorbed on a C18 column totally changing its mechanism. It is almost impossible to regenerate the column afterwards. Also as you are slowly adding SLS from the injection of diss. media it will never stabilize. Best approach is add SLS to the mobile phase [you will have to redevelop the chromatography]-this will solve the problem. But-put the column aside and use it only for this application. Also note equilibration time is about twice normal when you have SLS in the eluent."
I look forward for possible solution for such case.
I remember previously we developed a method containing Sodium octyl sulfonate as ion pairing agent and injected samples of 'same drug containg SLS' (in fact dissolution samples only). We had not observed any peak distortion in that case.
I look for the comments of the forum Members.