Gc Ms Voc BFB mass 50 fail too low and mass 174 to high
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
10 posts
Page 1 of 1
Hi, ive been having a problem with my gc ms with this masses around 4 days and no matter how many bfb tunes i used or how many times i use the tube wizard i cant manage to fix it, can someone help me. Thx
-
- Posts: 880
- Joined: Fri Jan 04, 2008 3:03 pm
More information please. What system are you using? Do you tune manually or use a "canned" tune?
I think this question has been discussed many times in the past so try searches for Method 524.
I think this question has been discussed many times in the past so try searches for Method 524.
-
- Posts: 4
- Joined: Fri Mar 18, 2016 2:13 am
Well i was using BFB tune and using tune wizard. But no matter how much i tried it always had a fail in mass 50. Whats a canned tune? Thnx in advance
-
- Posts: 880
- Joined: Fri Jan 04, 2008 3:03 pm
Sounds like you are using one. I use Agilent BFB_Autotune which you get with their water management kit for 5973 and 75's.
I have posted the instructions of how to do this tune manually in the past.
Please spend some time searching old posts.
I have posted the instructions of how to do this tune manually in the past.
Please spend some time searching old posts.
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
Definitely need to learn to manual tune the instrument if you have to pass BFB for your methods. Even the best autotune macro will sometimes not get it right.
What brand of GC/MS are you using? If we know that maybe someone will know more about how to help. But the main thing is you need to manually adjust the PFTBA masses so that you increase 50 versus 100 and reduce 131 and 219 versus 100, those are the nearest masses to the ones present in BFB.
What brand of GC/MS are you using? If we know that maybe someone will know more about how to help. But the main thing is you need to manually adjust the PFTBA masses so that you increase 50 versus 100 and reduce 131 and 219 versus 100, those are the nearest masses to the ones present in BFB.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 880
- Joined: Fri Jan 04, 2008 3:03 pm
James, did you mean m/z 69?
-
- Posts: 4
- Joined: Fri Mar 18, 2016 2:13 am
Well the ms is agilent, so u can do it manually? How can i do it manually? I just use autotune and BFB tune.The problem came out from nowhere. It had was functioning correctly, i had to use tune wizard and change the masses a little bit so the BFB could pass. And we had to change the source from another MS and that fixed the problem, but we still have to use tune wizard all the time.
-
- Posts: 880
- Joined: Fri Jan 04, 2008 3:03 pm
Ok, I feel your issue may also be due to all the water transferred to the ms during desorb. Please search this forum for dealing with this.
You may need a 6mm draw out plate in your source, helps to pump out the water faster.
Here is how I do a manual tune.
Run autotune ( save) also save as another unique name.
Set ion focus to 140 ( you may need to reset the upper limit in more functions)
set masses to 69,219, and 50
Scan and adjust ion focus till mass 50 is about 0.9
Set masses to 69,219, and 502
ramp emission and then set it to the peak of the scan for 69 ( just click on the graph on the screen)
set masses to 69,131,and 219
while scanning get 219 and 131 to about 50% of 69 ( use entrance lens and entrance lens offset)
Reset masses to 69,219, and 502
Scan and adjust the em to get ~ 600000 counts for 69.
Calibrate peak widths and mass axis
Check that 69 is still ~600000 cts
save tune values
if you are still having issues wit 174/176 this is the 131/219 in the manual tune (131 affects 174 while 219 does 176)
Good luck
Bear
You may need a 6mm draw out plate in your source, helps to pump out the water faster.
Here is how I do a manual tune.
Run autotune ( save) also save as another unique name.
Set ion focus to 140 ( you may need to reset the upper limit in more functions)
set masses to 69,219, and 50
Scan and adjust ion focus till mass 50 is about 0.9
Set masses to 69,219, and 502
ramp emission and then set it to the peak of the scan for 69 ( just click on the graph on the screen)
set masses to 69,131,and 219
while scanning get 219 and 131 to about 50% of 69 ( use entrance lens and entrance lens offset)
Reset masses to 69,219, and 502
Scan and adjust the em to get ~ 600000 counts for 69.
Calibrate peak widths and mass axis
Check that 69 is still ~600000 cts
save tune values
if you are still having issues wit 174/176 this is the 131/219 in the manual tune (131 affects 174 while 219 does 176)
Good luck
Bear
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
Actually there is a m/z 100 in the PFTBA and when I really have trouble getting m/z 50 high enough I focus on the 50/100 ratio in manual tune. If you set them and ramp the lenses you can see if there is one that will make a large difference with a small adjustment, which will help get 50/95 to pass in the BFB. Also if you can put 100 into the masses in the dynamic entrance lens offset table you can use that to help with the ratio, if not have to use 50/131 in the offset table to reduce the 100 response(95 in BFB). If you can get PFTBA 69/100 down to about 13% and 69/50 up to about 0.9% then it will almost always pass.James, did you mean m/z 69?
The past is there to guide us into the future, not to dwell in.
-
- Posts: 4
- Joined: Fri Mar 18, 2016 2:13 am
Thnx a lot that was the problem. The conditions from the purge and trap were incorect. This week came a tech an fix it. Thnx a lot for all the help
10 posts
Page 1 of 1
Who is online
In total there are 32 users online :: 0 registered, 0 hidden and 32 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 32 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 32 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science