Hello Everyone,

I am using a Waters UPLC for amino acid analysis with FLR detector. The middle peaks (between 3.5 and 6 min) started coming out after 4.5 min and now all "smooshed" together instead of equally spaced peaks. The peaks are stills separate enough to make out the individual amino acid but some can run together at the base. The peaks before 3 min and after 6 are coming out at the right times and look excellent.

I started seeing the problem every 10 samples or so while the majority of the sample came out correctly. Then my buffers ran low and I've made new buffers. The problem was then every sample. I made the buffers again. It worked for a few samples and then stopped (i would get the smooshed together middle peaks again).

The Waters trouble shooting guide suggested the column stabilizer. I fiddle around with it and standard came out correctly, then next standard did not and it hasn't worked since.

I've been checking flow rates through the entire BSM and can't seem to find where the flow is not correct.

After 3 min the Eluent B (ACN) starts to increase from 2% to 5 min 7% and the pressure increases a bit. At 5 minutes when most of the peaks come out (1 right after the next) the pressure increases again. Of course i'm not sure what the pressure should look like on a correct standard anymore.

So i don't know what to check anymore. Can anyone suggest anything?
thank you
In my mind the problem could be caused by different issues (not mutually exclusive).

1) Poor solvent preparation - maybe poor mixing?
2) inaccuracies in the solvent blending of the UPLC
3) degrading performance of the column due to wear

I would try to troubleshoot the problems in this order

regarding 1) the way to assess it is to prepare fresh solvents at least 3 times. Best way to make reproducible solvent blends is to measure them gravimetrically rather than volumetrically. You can put the cylinders on a balance and use density for calculations

regarding 2) every vendor will have its own methods, but this is a good starting point http://www.lcresources.com/tsbible/22102004.pdf

regarding 3) you should try a new guard column or a new column
2 posts Page 1 of 1

Who is online

In total there are 9 users online :: 3 registered, 0 hidden and 6 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot], Semrush [Bot] and 6 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry