Do they use the fixed 1ul rotor valve like the Archon did? If they are using the Valco valve as the 1ul loop for internal standard introduction, those will wear and allow standard to leak from the vial into the loop between samples and cause cross contamination. The only remedy is to replace the vespel rotor, and sometimes even the entire valve is something had happened to make the valve score the rotor.Has anyone else had problems with cross-contamination between on-board standards? I have 2 different Tekmar autosamplers: an Atomx and an Aquatek 100. I am having the same problem on both instruments. I'm now using only 1 channel for the IS/Surr Mix and doing everything else by hand because I can't figure out how to stop it.