lmh » Fri Jan 23, 2015 2:02 pm
(1) a lot of the issues that originally differentiated low-pressure mixing systems from high-pressure are a lot less relevant now, because the manufacturers have improved the design of both. Low-pressure mixing systems nowadays have much smaller dead-volumes than they used to.
(2) You have to differentiate between post-column and pre-column system volume. Post-column system-volume is a problem because it causes peak-dispersion and it's an unavoidable delay. Pre-column is less serious because there is no peak to disperse at that point! It delays the arrival of the gradient at the column (which is something for which some software can compensate, by starting the gradient early) which makes runs longer (bad). It also acts as an extra mixing-volume, potentially smoothing sudden changes in solvent composition - but that is probably desirable!
(3) The change in volume on mixing methanol and water is pretty small. I've seen 4% bandied about, but only in unreliable forums. I suppose it's also possible that the change in temperature is relevant if you're worried about bubbles appearing in the mixing chamber of a low-pressure system (increased temperature in MeOH/water mixing should reduce solubility of dissolved air). The real issue about bubbles forming in a low-pressure mixer is what happens in the pump heads after they've formed. Some systems tend to accumulate bubbles in the pump-head, which is disastrous. Others are more tolerant, and flush them straight through into the high-pressure part. If this happens, the bubble will vanish and redissolve as soon as the pump piston starts moving, and the overall effect will be a 4% loss of flow, and possibly some pressure fluctuation depending on how well the pulse dampener is doing its job. Note that a high-pressure mixing system would automatically have a 4% reduction in flow when creating the same % mix, because the mixing and reduction happen after the pistons that do the measuring. A low pressure system starts off with 0% reduction because the pistons are working with pre-mixed, "pre-shrunk" mixture. As Klaus I. said, this is probably irrelevant in practice. Degassing is, in any case, a jolly good idea.
(4) Pumping low percentages in high-pressure mixing systems. I don't think this is (usually) a problem. What doesn't work is pumping a gradient from 0-3%, particularly at low flow. Basically it's problematic because the step-size of the pump delivering the low-percentage solvent becomes large compared to the overall gradient.
What you have to think about is this: a high-pressure pump can deliver a range of flow-rates in definite steps. The steps are usually very small, but imagine a (rather bad!) pump that delivers 0-1mL/min in 1uL/min steps. If you use it in a high-pressure mixing system, and pump an overall flow of 100uL/min, and try to pump a gradient from 0-10%, the pump doing the low-side of the gradient has to start at 0uL/min and progress linearly to 10uL/min, but it can only do this in 1uL/min steps, so the gradient isn't linear; it's a staircase of 10 steps, and if you do it over 20 minutes, it means you're not really running a gradient, you're doing 10 successive 2-minute isocratic runs!
It's OK to get your pump to start at 0% provided you are working at a flow-rate, and with a gradient, where the lowest flow the pump can actually deliver, and the step-size by which it sets its flow, are too small to matter. You can't start at 0% and run a gradient if that means that your pump is operating at its minimum flow of, say, 0.1uL/min, and the elution of your favourite analyte is critically dependent on this being 0.1uL/min and not 0.2uL/min! But if that is the case, you are in any case stuck, because starting at a higher percentage won't help.
I suppose what I'm getting at is that you can pump a gradient from 0%, but it will actually be, maybe, 0% +/- 0.2%. You could also start at 1% +/- 0.2%, which would look like a 20% variability on the actual pumped mixture (proportionally speaking), which looks bad. You could start at 5% +/- 0.2%, which is a 4% variability in what's actually pumped, which looks a lot better. But in any properly designed gradient run, an actual, absolute variation of 0.2% in the absolute mix is irrelevant.
Hope this makes some sense???