by CE Instruments » Mon Nov 17, 2014 4:47 pm
													
						
					 
					
						I've tried to Google a bit and this is what I've got:
1. Porapak Q Column (actually, I thought it was an Hayesep Q, but I just checked and it seems I was wrong) - No information about conditioning - just 250°C Maximum Temperature
2. Restek Rt-Q-Bond 300°C Maximum Temperature - Restek suggest a 4h 200°C isotherm for conditioning
3. Molecular Sieve 5A - Conditioning: 300°C for 3h
Why do you think you need to condition the columns ?  Are they new ?  Are they all installed in one GC ?  Do you have a gas analyser?
The easiest way to condition a GC is to leave it cycling over night running your temperature program.  Unless you have a specific issue with your separation I would just stick to this.
No, they're not new but, correct me if I'm wrong, I thought tha conditioning was a standard procedure for "restarting" a GC that has been shut down for a long time (2 years); also, we're not really pleased by our analysis reproducibilty (
not sure if could be a column or maybe a detector problem - all suggestions are welcome). Yes, they're all installed on 1 GC. What do you exactly mean by "gas analyser"? Are you talking about GC detector? Sorry if you find this question stupid 
Actually, what we will try is to set an isotherm at 200°C (the lowest maximum temperature suggested for conditioning), and try to recondition all the columns together[/quote]
 
  
  
 
Now we know what columns you have and that they are installed in one GC it is clear you have some sort of permanent gas analyser  
 
If the GC has sat for a couple of years you could well have separation issues some or all of which may be sorted by leaving the GC at 200C for a while however without a flow diagram we do not know if carrier flows through all the columns when left iso thermal.  
The more info. you give us the less "stupid"  

  the questions and the more useful the answers will be