MS does not activate in Agilent 5973N GCMS
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
4 posts
Page 1 of 1
We changed a column on our GCMS and when it was restarted, the instrument will not collect MS data, as if the detector function is not working. The instrument will go through a normal autotune, so we know the filament and source is working. The instrument will begin a run as it would normally, though it does not prompt with the question "Override solvent delay at 4 minutes?" and there is no data from the MS detector. All of our MSD methods were written about 10 years ago, so no one knows how to check the methods to control MS functions, or if another type of issue might be causing the problem. It seems likely to be a software problem - or user problem - more than a hardware problem. Any suggestions? Thanks.
-
- Posts: 7
- Joined: Fri Feb 24, 2012 9:49 am
Normal problem with these - just power cycle the MS and it will all run again!
-
- Posts: 526
- Joined: Fri Feb 10, 2012 7:39 pm
There's a little box ("Edit Entire Method") which needs to be checked, will say "Use MSD" next to it.
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
This could be it. Using Edit Entire Method it will be on the second screen that appears where you will find the check box.There's a little box ("Edit Entire Method") which needs to be checked, will say "Use MSD" next to it.
The past is there to guide us into the future, not to dwell in.
4 posts
Page 1 of 1
Who is online
In total there are 8 users online :: 0 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 8 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 8 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science