Late negative peaks found by using HPLC with RI detector
Posted: Thu Jan 22, 2015 8:45 am
Hi all
Currently i am testing USP monograph, Mannitol, related substance:
USP parameters are as below:
Mobile phase: Degassed water
Mode: LC
Detector: Refractive index
Column: 7.8-mm × 30-cm; packing L19
Temperatures
Column: 85 ± 2
Detector: 40 (maintain at a constant temperature)
Flow rate: 0.5 mL/min
Injection volume: 20 µL
Run time: NLT 1.5 times the retention time of the Mannitol peak. [Note—The retention time for Mannitol is about 20 min. ]
The HPLC I used is Waters e2695 with 2414 RI detector (corrected). The column I used is the same type L19 and the same length (Phenomenex RCM). Rest of parameters follow USP monograph. MP and needle wash is water (Merck, for chromatography). Diluent is the same solvent, which means that the refractive index should be the same with m.p. However, after the injection of diluent, 3 negative peaks appeared in the region where the Known Impurities elute at about 18 min. Even I inject 0mL of diluent, 3 smaller negative peaks still appeared in the same region.
At first, I thought the different external volume might cause those peaks shift because those negative peaks can be separated from Impurities by using the same Instrument and column about one year ago. So I changed the different tubing length between HPLC and column. However, the retention time didn’t change at all.
Therefore, I assume that there might be something wrong with the water to cause negative peaks (Merck, for chromatography. Same situation occurred for in-house purified water) regarding the result of 0mL of diluent (I guess that the residual water adhered to the outside of syringe was brought into the system while injecting). But I could not explain why there are “late” and “negative” peaks as diluent is the same solvent.
Hope you could provide me any valuable suggestion to solve my problem. Thank you in advance!
Currently i am testing USP monograph, Mannitol, related substance:
USP parameters are as below:
Mobile phase: Degassed water
Mode: LC
Detector: Refractive index
Column: 7.8-mm × 30-cm; packing L19
Temperatures
Column: 85 ± 2
Detector: 40 (maintain at a constant temperature)
Flow rate: 0.5 mL/min
Injection volume: 20 µL
Run time: NLT 1.5 times the retention time of the Mannitol peak. [Note—The retention time for Mannitol is about 20 min. ]
The HPLC I used is Waters e2695 with 2414 RI detector (corrected). The column I used is the same type L19 and the same length (Phenomenex RCM). Rest of parameters follow USP monograph. MP and needle wash is water (Merck, for chromatography). Diluent is the same solvent, which means that the refractive index should be the same with m.p. However, after the injection of diluent, 3 negative peaks appeared in the region where the Known Impurities elute at about 18 min. Even I inject 0mL of diluent, 3 smaller negative peaks still appeared in the same region.
At first, I thought the different external volume might cause those peaks shift because those negative peaks can be separated from Impurities by using the same Instrument and column about one year ago. So I changed the different tubing length between HPLC and column. However, the retention time didn’t change at all.
Therefore, I assume that there might be something wrong with the water to cause negative peaks (Merck, for chromatography. Same situation occurred for in-house purified water) regarding the result of 0mL of diluent (I guess that the residual water adhered to the outside of syringe was brought into the system while injecting). But I could not explain why there are “late” and “negative” peaks as diluent is the same solvent.
Hope you could provide me any valuable suggestion to solve my problem. Thank you in advance!