Replacement Pump for Agilent 1322a degasser

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Greetings all! I isolated a problem on my Agilent 1322a degasser down to a worn pump. I was wondering if anyone had suggestions for a replacement pump that would fit without costing an arm and a leg.

Thanks

Casey
It's basically an aquarium pump (diaphragm pump) hooked up in reverse.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Speaking as a representative of a third-party instrument service provider, there's no easy way of getting around Agilent as the source of a replacement pump that will work in that degasser. Make sure the vacuum tubing between diaphragms is okay, and if you feel industrious you can try disassembling the pump and cleaning the diaphragms, but beyond that the best option is to replace the vacuum pump (P/N G1322-6000).

-dhobbs
Thanks for the advice all.

I ended up taking it apart, cleaning it and putting it back together and it is working pretty well so far.

Casey
4 posts Page 1 of 1

Who is online

In total there are 25 users online :: 1 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: John Guajardo and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry