Diisononyl Phthalate (DINP) Gas Chromatography
Posted: Thu Jan 22, 2015 3:14 am
Hello everyone,
I took over a GC analysis for phthalate testing from a co-worker, and I am completely lost over what I'm doing right or wrong. I am unfamiliar with the capillary column, as I work mostly with water solvents and extracts thus packed column is my forte.
Per the recommendation of a test method, these are the parameters of the Agilent 6890N I'm using right now to run 0.05 ug/mL, 0.5 ug/mL and 5.0 ug/mL of DINP standards:
Instrument: Agilent 6890N
Column: Chrompack fused silica, CP-Sil-5CB, 50 m x 0.32 mm, 0.12 um id
Temperature Program: Start at 150°C, ramp 5°C/min to 280°C, hold at 280°C for 5 minutes
Carrier gas: Helium at 44 mL/min, column head pressure: 19.5 psi
Injector: Split, 300°C, ratio 20:1
Detector: Flame Ionization, 300°C
Make-up Gas: Nitrogen at 30 mL/min
I am using a brand new column which has already been conditioned for 2 days, and am currently running more blank solvents and standards to determine the DINP peak from the chromatograms. I am on a very tight deadline, any input at this point will definitely help. I'm also using the Agilent guide on a method that was performed per the Chinese standards on phthalate testing on children's toys, if anyone can recommend any other tweaks to the methods.
Your efforts are truly appreciated!
I took over a GC analysis for phthalate testing from a co-worker, and I am completely lost over what I'm doing right or wrong. I am unfamiliar with the capillary column, as I work mostly with water solvents and extracts thus packed column is my forte.
Per the recommendation of a test method, these are the parameters of the Agilent 6890N I'm using right now to run 0.05 ug/mL, 0.5 ug/mL and 5.0 ug/mL of DINP standards:
Instrument: Agilent 6890N
Column: Chrompack fused silica, CP-Sil-5CB, 50 m x 0.32 mm, 0.12 um id
Temperature Program: Start at 150°C, ramp 5°C/min to 280°C, hold at 280°C for 5 minutes
Carrier gas: Helium at 44 mL/min, column head pressure: 19.5 psi
Injector: Split, 300°C, ratio 20:1
Detector: Flame Ionization, 300°C
Make-up Gas: Nitrogen at 30 mL/min
I am using a brand new column which has already been conditioned for 2 days, and am currently running more blank solvents and standards to determine the DINP peak from the chromatograms. I am on a very tight deadline, any input at this point will definitely help. I'm also using the Agilent guide on a method that was performed per the Chinese standards on phthalate testing on children's toys, if anyone can recommend any other tweaks to the methods.
Your efforts are truly appreciated!