megabor installation on the agilent 6890N
Posted: Tue Jan 20, 2015 3:58 pm
Do I need an adaptor to install a megabor column on the 6890N GC? It is currently set up to run a capillary column.
Chromatography Forum is a public discussion group where you can post questions, news, or messages of interest to chromatographers everywhere, but you must be registered to participate (registration is FREE). If you are a registered user, please log in.
http://www.chromforum.org/
Hi CPG, the larger holes create more of problem with graphite ferrules extruding past the column if you use them with thin columns.It "can" use the same nuts, but of course different ferrules.
I don't know what is sold these days, but in the past there were nuts for smaller size capillaries that actually had a smaller hole than for the 0.53mm columns. The nuts may all be supplied with the larger hole these days, can't imagine why those nuts with the larger hole shouldn't work for all common sizes.