Empower Noise/Drift calculation for Shimadzu UV detector

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

2 posts Page 1 of 1
Hello All,

I'm doing a system validation for a Shimadzu UV detector with Empower. The validation includes also the noise/drift test. These test have a criteria of 6.0x10-e4AU for Drift and 2.0x10-e5AU for Noise.

From the Empower i have only Plots/units and not AU units. Can someone help me of how can i convert these units?
Or maybe there is someone that had a simliar problem when validate a Shimadzu with Empower/

Thanks for your suggestions
Do you use a SAT/IN box?

In the instrument method you can change the stettings of a sat/in box. In here you can change the unit in a plot I.e. in AU.