HPLC validation - accuracy determination issue

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
A colleague needs to determine accuracy (impurity level) of a method but the material has a high impurity content which will confound the true recovery. Said colleague states that because the impurities and parent compound are well separated by the chromatography, the parent will have no impact on the impurity analysis. Thus, it is not necessary to have the parent material present in the impurity validation test solutions, i.e., neat solutions not spiked solutions.

Pretty sure this is not ICH compliance, but would like input from others..

How would one solve a high impurity content issue?

regards,
Standard addition quantitation.
Std addition works okay for low impurities levels in the spike std. However, if the impurities are above the LOQ in the spiked std, then you may want to consider spiking the placebo with a lower std conc (where the impurities is significantly reduced) plus the neat impurities. Spiked neat impurities only in placebo should be your last resort.
3 posts Page 1 of 1

Who is online

In total there are 590 users online :: 1 registered, 0 hidden and 589 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 589 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry