PhD Candidate, Mentor left, pressure issues
Posted: Wed Jan 14, 2015 3:31 pm
Hi there,
I'm running out of resources here. Its my final year and my PI left for a different job. I'm now alone in the lab for these remaining months trying to keep it going with running the rest of the grants and company contracts. I need to get the HPLC running nicely for my last group of experiments, but since my mentor has gone, I have nowhere to turn to figure out how to fix these problems. Not sure if this is the right place to ask... but, if not, could you kindly point me in the right place?
the issue...
I'm running a microbore column to detect adenosine from microdialysis samples. Right now, I'm just trying to get standards going. Its a UV detector with a manual injector. The issue is pressure.
The column is a few years old, but I just put it into use two weeks ago (I have run it non-stop since, only having maybe a day or two where I wasn't running standards and calibrating the microdialysis probe head. I am under the impression keeping it running in 50:50 wash buffer is the best thing to do instead of letting it sit unused and not flowing at any point in time. Maybe I'm wrong about this?) and at first, my pressure with 50:50 MeOH:H2O was 15MPa. I have only run standards through this thing to calibrate the microdialysis probe, so no samples from the rat were run. Just standards in polished water.
Last week, the pressure dropped from 15MPa to 11MPa overnight. I figured out that the pump's plunger seal needed to be replaced so I ordered it. The next day, after running the system overnight to wash, the pressure was back up to 15MPa. I ran some standards again that day. The next day, the pressure was all the way up to 25MPa!!! I changed the wash buffer in case it was contaminated (degassed and filtered). I replaced the plunger seal and confirmed that it was working correctly. I completely cleaned out the manual injector by running wash buffer through the sample resevoir loop and pushing wash buffer through the injector port while in the "trigger" position. I did this while the column was off to make sure it didn't hit the column.
I reversed flushed the column for about two hours. When I put the column back in position, it ran at only slightly above normal pressures for hours.
Then, overnight, it jumped all the way up to 30MPa again!!!!! This time, it hit this pressure at a flow rate of 30ul/min. The above pressures were for flow rates of 50ul/min.
I have tested the column. The standards work. I can detect adenosine, but this pressure is getting WAY too high. Something is wrong and I have nowhere to turn to try to figure this out, and I have little money left to call out a tech support for a visit.
Since the pressure climbs overnight, I almost feel like it is something that is aggregating over time that builds up on the column entry point and raises the pressure. I think this because reverse flushing seems to fix it. I have a guard column with a new filter on it too and have always used this. I have heard that some parts of the system can sometimes wear away and cause clogs. What parts of the system do this? What do you think I should replace? This overnight pressure build up makes me think this.
Could it be worn seals somewhere I'm unfamiliar with? Worn tubing? Worn filters?
I should also mention that for nearly two weeks, I would run this column every night non-stop at a low flow rate with wash buffer. It never did this. It wasn't "slowly" getting worst and worst. It seems like in a single day, it just instantly developed this "overnight pressure" issue. Like, BAM. Here it is... every night, it will do this. I just started revrese flushing it again, and at the same flow rate... its all the way to 16MPa now. That actually speaks against it being an issue of something wearing apart and catching debris since if it was wearing apart, I would expect it to wear apart slowly...slowly developing a higher pressure each day...instead of all of a sudden, developing this issue in a single night, then doing the same every night. BUT, when I reverse flush it, it seems to fix the problem, suggesting I have pushed off any debris that has built up.
Ugh...it sucks not having a PI here, who has all this experience with this incredibly old machine, to help me fix this situation.
Thank you.
the issue...
Thank you.