Latest USP<621> guidelines re method adjustment HPLC / UHPLC
Posted: Fri Jan 09, 2015 12:11 pm
I have a project for adjusting some USP methods from "HPLC" to "UHPLC", i.e. smaller particle size, shorter columns, faster separation. I thought I would just check the latest USP Chapter 621 guidelines (released August 1, 2014) as, like many others, I have been waiting for the updates that would allow more flexibility in choice of column dimensions and particle size.
I was surprised to see the statement that says: "For gradient separations, changes in length, column inner diameter and particle size are not allowed".
This means we can only change isocratic methods and yet I know from experience that gradient methods can be successfully and easily converted when the gradient steepness based on column volumes is kept the same for each gradient step and making an allowance for difference in delay volumes between systems.
Has anyone else got views or comments on this?
I was surprised to see the statement that says: "For gradient separations, changes in length, column inner diameter and particle size are not allowed".
This means we can only change isocratic methods and yet I know from experience that gradient methods can be successfully and easily converted when the gradient steepness based on column volumes is kept the same for each gradient step and making an allowance for difference in delay volumes between systems.
Has anyone else got views or comments on this?