- Posts: 2
- Joined: Fri Feb 20, 2015 7:14 pm
I am having an unusual problem with back pressure. My machine, an older Agilent 1100, is used heavily. I installed a new column (RP-18) and after only four or five days of use, the back pressure was way up (should be around 28bar, climbed to 50bar). The rest of the system has pressure that is just a little elevated (6bar or so) so I'm thinking that I have some seals or something breaking down that are contributing to small particles in the system and prematurely ruining my column. I installed another new column to test this and sure enough the pressure stabilized around 28 and a few hours later is up to 35. My mobile phase is isocratic methanol.
My real question: I have read the relevant trouble shooting guides, but I am looking for the opinion of an experienced user. Which seals should I replace first? I replaced the needle seat assembly and needle recently, and I do the purge valve PTFE frit regularly. Everything else has been in place for at least 1-2 years (I inherited this instrument in late 2012).