ICS-3000 Ion Chromatography bubbles
I'm new to IC but the system seem to have a constant problems with bubbles that seem to worsen lately. We use the IC to test water samples only (6 anions and by products). The eluent is refilled on a daily bases but this cause bubbles to form, we prime the pump. This causes the peaks of samples to shift from the calibration peaks (retention time changes), we run the standards again to re-do the calibration. We have to re-do the calibration 3-4 time a month (only this January we did it 3 times so far). Is this normal?
How do you degass your mobile phase???????
Hello
I'd check:
-mobile phase filter (replace or clean) - air bubble can be from filter
-degasser connections - check if all are tight enough (air can be sucked in there)
Regards
Tomasz Kubowicz
They used to use N2 for degassing, but stopped it recently.
Also, the stopped using the filter because it was causing even more bubbles.
Can the priming cause the retention time to change?
Air bubbles in the mobile phase, pump can influence RT. can you ultrasonic your mobile phase when you start in the morning?
You're joking right? Air is 80% N2, so bubbling N2 through the mobile phase is the exact opposite of what you need (no wonder you had bubble problems!). Degassing can be done with helium, because He is almost insoluble.
Back in the days of old with the Dionex DX-400, we boiled the conductivity water and allowed it to cool under Ascarite before making IC eluents...that was obviously QUITE a while ago. Did a good job with the retention times of analytes (though it was a true bother), not so good with background conductivity, but then I wasn't doing a whole lot of trace-level work.
He is an excellent sparging gas...when I was asked to try nitrogen for HPLC work, I found that the pumps "choked" pretty often...my question here is, is sonication of the water in conjunction with applied vacuum a possibility...prior to putting the degassed water into a bottle to feed the eluent generator?
My thanks.
MattM
In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)