Hello all,

I inherited two older Shimadzu LC 10 ADvp pumps and want to perform a preventive maintenance. Unfortunately I don`t have a manual or service manual.
Changing the pistons, seals and valves are no problem. But I am not sure about the diaphragms. Must they be changed when I change the pistons or seals? Can they be reused ?

I would appreciate if somebody has an manual or service manual for the LC-10ADvp.

Thank you very much in advance.

Stephan
The diaphragms are for automatic seal washing , they circulate water at the back of pistons , and if they are in good conditions( no cracks or deformations ) no need to change.
Dear Uzman,

thank you very much for your answer. I am very glad to hear this.

best regards

Stephan
however, those parts are relatively expensive. They should be changed with seals especially if you're changing out everything else. You can just go to the shimadzu website to order them.

The nut of the piston keeps that in place.
4 posts Page 1 of 1

Who is online

In total there are 12 users online :: 1 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry