On a chromotogram, how come there is a baseline negative excursion before the first anion peak? (we were using IonPac AS 14 column for anion exchange.3.5 mM NaCO3/1.0 mM NaHCO3 buffer as mobile phase) I've searched in many texts but just can't find the answer. Is there any explanation for the excursion showing up sometimes and sometimes not, and sometimes it's a large negative one, but sometimes just small? Your help is very much appreciated!!
snowysummer,

The negative excursion you observe at the beginning of the chromatograms is a direct consequence of the fact that the background you are working from when using the standard eluent system is due to the presence of 4.5 mM carbonic acid (the byproduct of the suppression reaction). When you inject a sample containing no carbonate, there is a diminished background signal corresponding to the system void volume due to the fact that the sample contained little or no carbonate. If you inject eluent, you will see no disturbance is located at the system void volume. If you inject a double concentration solution of eluent you will see a peak at the system void volume location. It might seem confusing as to why you observe this disturbance at the void volume considering the fact that the retention time for carbonate is significantly different than the time for this baseline disturbance to manifest but this is a direct consequence of the fact that stoichiometry must be maintained for anions to relative to cations as the sample passes through the column. Consequently, there is a disturbance corresponding at the system void volume with an ionic strength exactly the same as the ionic strength of the sample that was injected. If you studied the situation further with radiotracers you can show that all analytes, including carbonate present in the sample are eluted normally as one would exact in an ion exchange process. The carbonate present in the sample in such a case actually elutes at the retention time one would normally expect for carbonate. The disturbance at the system void volume involves carbonate present in the eluent not carbonate present in the sample. The situation is actually a bit more complex than this due to the fact that you are using a buffered system but the above explains the basic cause of the negative excursion.
thank u so much for ur reply!!!
3 posts Page 1 of 1

Who is online

In total there are 87 users online :: 1 registered, 0 hidden and 86 guests (based on users active over the past 5 minutes)
Most users ever online was 11068 on Sat Dec 06, 2025 9:55 pm

Users browsing this forum: Google [Bot] and 86 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry