Triplus autosampler problems
Posted: Mon Dec 15, 2014 11:47 am
by s.eaton@ucl.ac.uk
I have a Thermo Triplus autosampler attached to a Trace GC with DSQ II MS. The liquid head does not initialize properly any more - random movements, whereas the headspace autosampler head initializes fine. There are no serviceable parts to this liquid head and a new head is very expensive. Any ideas what might be causing the problem?
Thanks,
Simon
Re: Triplus autosampler problems
Posted: Mon Dec 15, 2014 2:49 pm
by CE Instruments
Bad or dirty sensor on the liquid injection head. There may be no user serviceable parts on the sampler but I would be surprised if it was not servicable.
Re: Triplus autosampler problems
Posted: Mon Dec 15, 2014 2:52 pm
by s.eaton@ucl.ac.uk
That's what I thought, it can't be completely trashed. Thermo do not offer parts, just a new turret. Will try cleaning the sensor as you suggest.