Hi guys--

How long do your standards (IS and SS) last in the vials on your autosampler? I replaced one of the vials and refilled them and since then, they have been evaporating pretty quickly (half a mil per week, maybe?) This seems too fast, and I wonder if I've installed the vials incorrectly. I'm new at this and have had to train myself. Any info you can give would be super. Thanks!
There is a teflon ferrule that makes the seal at the top of the vial. Replace it frequently. Also check where the tubing enters the top of the vial holder.
In my case the fluorobenzene loss in the headspace makes the 1,2-dichloroethane-d4 recovery fall out of spec. It has been as bad as 1 day, fixed with a new valve rotor, and as good as >1 week.
Ahh, thank you! I may not have had the reservoir tightened up against the ferrule all the way. I noticed the problem all of a sudden when I removed the vial for filling. If this doesn't help, I'll replace the ferrule. Thanks again!!
3 posts Page 1 of 1

Who is online

In total there are 202 users online :: 0 registered, 0 hidden and 202 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 202 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry