How to analyze dissolved N2O in water with GC-ECD

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Hello,
I am trying to measure N2O Using GC-ECD with PLOT Q column without head-space arrangement.
I need to measure in aqueous phase.
Kindly suggest how I can go about.
Hello akellacs,
Maybe more info from you would help. So you do not have "headspace arrangement". How are you delivering samples to the GC, ie injection? The most common way to measure dissolved N2O (in water, seawater, slurry) is by headspace. So if you have a water sample, a headspace is introduced (high purity helium for example). The sample is then allowed to equilibrate (at known temp, shaking helps) between gas and water phase. The gas phase N2O concentration is analysed. With known solubility you can calculate dissolved N2O concentration in your samples. We do this occasionally for seawater, sediment slurry samples. Also there are N2O electrodes out there, that are very sensitive and could be used in situ.

Cheers, Nick
2 posts Page 1 of 1

Who is online

In total there are 202 users online :: 0 registered, 0 hidden and 202 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 202 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry