I have a problem with an air peak using SPME. The air peak appears at the beginning of the chromatogram. It’s air as I checked it with scan range 18-100 m/z.
The peak doesn’t appear due to the leak in the injector port as when a GC programme is run (with no fiber introduction), there is no air peak. It does appear when the SPME fiber is introduced into the injector port. The longer the SPME fiber in the injector port, the more intense is the air peak. The SPME fibers are the right size (23 ga) and I used a few (the peak is always there).
My SPME Merlin Microseal is new. But I also checked it with another, older (~ 200 injections) SPME Merlin Microseal and it doesn’t make any difference. I thought as Merlin claims that ‘septum leaks are acceptable as long as system pressure is maintained’, I will check it. I have CP-3800 Varian GC equipped with the 1079 injector port and EFC Type 1. I have calibrated the septum purge but the peak is still there. When I change the gas flow from 1 ml/min to 2 ml/min, it does not make any difference.
What is more, I checked some old SPME experiments (March 2013, I had a break with it for a while) and this peak WAS present then, however it wasn’t very intense and I ignored it as a fiber contamination. It looks like the problem is getting worse. I cannot ignore it now as the air is stripping off the siloxanes from my fibers. They may last shorter and the siloxanes are making my chromatograms messy.
If it’s not a leak in the injector port, a problem with septum purge or a faulty Microseal, I am running out of ideas. I have calibrated the septum purge at 4.5 ml/min (column pressure 4.8 psi for 1 ml/min and oven turned off).
Any advice will be very appreciated,
Thank you,
Kamila