Planned maintenance for Shimadzu LC10 ADvp
Posted: Tue Dec 09, 2014 8:30 pm
by labstephan
Hello all,
I inherited two older Shimadzu LC 10 ADvp pumps and want to perform a preventive maintenance. Unfortunately I don`t have a manual or service manual.
Changing the pistons, seals and valves are no problem. But I am not sure about the diaphragms. Must they be changed when I change the pistons or seals? Can they be reused ?
I would appreciate if somebody has an manual or service manual for the LC-10ADvp.
Thank you very much in advance.
Stephan
Re: Planned maintenance for Shimadzu LC10 ADvp
Posted: Tue Dec 09, 2014 9:08 pm
by uzman
The diaphragms are for automatic seal washing , they circulate water at the back of pistons , and if they are in good conditions( no cracks or deformations ) no need to change.
Posted: Fri Dec 12, 2014 3:35 pm
by labstephan
Dear Uzman,
thank you very much for your answer. I am very glad to hear this.
best regards
Stephan
Posted: Sun Jan 25, 2015 7:34 am
by yazmer
however, those parts are relatively expensive. They should be changed with seals especially if you're changing out everything else. You can just go to the shimadzu website to order them.
The nut of the piston keeps that in place.