FID air ratio nitrogen/oxygen

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Dear all
I have some experience in chromatography but I am an absolute beginner concerning classic detectors like the FID. (I did a lot of stable isotope work using IRMS in the last years). Now I am in the position to re-establish a lab to measure CH4 in lowest amounts (4 mL of air STP, lowest concentrations: 350 ppb CH4).

What I found in that lab is a bottle with N2/O2 = 60/40. From my manual I expect AIR = N2/O2 = 80/20.
Furthermore I reconstructed that my forerunners (not available) used an old Varian GC (approx. late 1980s) with the following flow rates for the FID:
H2 = 30 mL/min
60/40 N2/O2 = 330 mL/min
N2 = 20 mL/min
Carrier = He = 5-10 mL/min

I assume N2 was used as make up gas.

Can anyone explain, why someone would use N2/O2 = 60/40?
Could it be that this increases sensitivity for lowest amounts of CH4?
Could it fade out a disturbing species (like CO)?
Could it be that N2 was not used as make up but to dilute the 60/40 mix?

Any hints are welcome. My only knowledge on this is from textbooks and the manual (saying: air, no specs on N2/O2 ratio). Please forgive if essential info is missing here, please tell me what I forgot.

My new GC will be a: Agilent 7890B equipped with a FID for capillary columns only.

Thanks!
Michael
:shock: Never heard of anyone using a 60/40 mix. FIDs usually work best with a slight increase in Nitrogen over the standard Air mix, see Special Gases "FID Air" specifications. Your new GC will come with recommended settings including a different set of values for ignition controlled by the GC. If you can afford a new 7890 then it is probably sensible to junk the 60/40 mix and get some standard air.
Thanks CE Instruments

Money for the air bottle is not an issue. Actually, if I go for 80/20, I might use a zero air generator that is not needed by a colleague.

You state
"FIDs usually work best with a slight increase in Nitrogen over the standard Air mix"
.
The question is what is the "usual" application? I speculate that the specific needs here for lowest trace amounts of CH4 was the reason to go for 60/40 (N2/O2).
However, if the latter is something odd, and no more information is around, clearly I will start with the usual 80/20 mix and see how far I come in terms of sensitivity...

Best
Michael
3 posts Page 1 of 1

Who is online

In total there are 207 users online :: 0 registered, 0 hidden and 207 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 207 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry