USP monograph link: http://www.usp.org/sites/default/files/ ... nnitol.pdf
Currently i found that the content of related substance of Mannitol monograph has been revised per USP37 2nd supplement and I have few questions as follows:
1. Detector temp: It is set to "40C (maintain at a constant temp)". In previous version, it only stated "maintain at a constant temp". In our previous verifcation, it was set to 50C. In this case, should I re-verify our method as I can not find any modification range of RI detector temp regarding USP37 <621>?
2. As this related substance is limit test, which solution listed in the monograph (System A, System B, STD A, STD B and STD C) should be the standard solution to check HPLC system suitability (%RSD, n=6)?
In my opinion, System A and System B is to check resolution and identification, STD A is for Assay purpose, STD B (maximum limit) is compared by Sample, and STD C is as sensitivity solution. So in this case, STD B should be the STD solution to check HPLC system suitability.
Hope you can provide me any valuable information, reference or weblink. thank you in advance.!