Refractive Index Detector + Acetonitrile = noise!
Posted: Wed Nov 26, 2014 6:04 pm
Hello all,
I'm completely new to the forum and I mostly registered because I realised that the client support and lab technicians I work with are not real experts concerning my system.
My goal is to separate and quantify different monosaccharides and for that I use the following setup:
- All hardware except for the tubing and the column is from the Agilent 1260 Infinity series
- Iso quartenary pump G1310B (includes degasser)
- ALS (autosampler) G1329B
- TCC (column oven) G1316A
- RID (refractive index detector) G1362A
I tried three columns:
- Shodex SP0810 + guard column
Optimised the separation at 86/14 V% Water/Acetonitrile (premixed, isocratic), flow = 1ml/min, Tcolumn= 80 C, Trid = 55 C, injection 10 ul, 20 min. run
- Bio-rad HPX-87H
Optimised the separation at 5 mM H2SO4 in Water (premixed, isocratic), flow = 0.6 ml/min, Tcolumn= 65 C, Trid = 55 C, injection 10 ul, 20 min. run
- Zorbax NH2
Tried 80/20 V% Acetonitrile/water (premixed, isocratic), flow = 1ml/min, Tcolumn= 40 C, Trid = 40 C, injection 10 ul, 20 min. run
The Shodex gave the best results in terms of separation, but the solvent peak interferes with the glucose peak, the baseline is really shaky in the low concentration range.
The Bio-rad could not separate the sugars completely, only 5 of the 8 sugars were clean peaks (but then it is not a specialized sugar column)
The Zorbax NH2 I could not check because of a bad baseline.
In all cases the pump pressure does not have a bigger deviation than +/- 0.4 bar. Everything without acetonitrile gives a slick, flat, clean baseline, but as soon as I enter acetonitrile in the system (premixed, isocratic!) the baseline starts jumping like crazy! I cannot distinghuish concentrations of 0.5 mg/ml from the baseline. This is after leaving the system flush for > 2h and flushign the purge cell of the RID > 30 minutes.
I know that the difference in viscosity and refractive index can give problems, but I thought that if the solvent is properly mixed and degassed that should not happen anymore.
If that helps I can post some spectra with and without injection for the Zorbax (since I use the highest acetonitrile concentration there).
Hopefully I am just inexperienced and this matter can be solved very easily, I'm looking forward to your comments!
Best regards,
Adriaan
I'm completely new to the forum and I mostly registered because I realised that the client support and lab technicians I work with are not real experts concerning my system.
My goal is to separate and quantify different monosaccharides and for that I use the following setup:
- All hardware except for the tubing and the column is from the Agilent 1260 Infinity series
- Iso quartenary pump G1310B (includes degasser)
- ALS (autosampler) G1329B
- TCC (column oven) G1316A
- RID (refractive index detector) G1362A
I tried three columns:
- Shodex SP0810 + guard column
Optimised the separation at 86/14 V% Water/Acetonitrile (premixed, isocratic), flow = 1ml/min, Tcolumn= 80 C, Trid = 55 C, injection 10 ul, 20 min. run
- Bio-rad HPX-87H
Optimised the separation at 5 mM H2SO4 in Water (premixed, isocratic), flow = 0.6 ml/min, Tcolumn= 65 C, Trid = 55 C, injection 10 ul, 20 min. run
- Zorbax NH2
Tried 80/20 V% Acetonitrile/water (premixed, isocratic), flow = 1ml/min, Tcolumn= 40 C, Trid = 40 C, injection 10 ul, 20 min. run
The Shodex gave the best results in terms of separation, but the solvent peak interferes with the glucose peak, the baseline is really shaky in the low concentration range.
The Bio-rad could not separate the sugars completely, only 5 of the 8 sugars were clean peaks (but then it is not a specialized sugar column)
The Zorbax NH2 I could not check because of a bad baseline.
In all cases the pump pressure does not have a bigger deviation than +/- 0.4 bar. Everything without acetonitrile gives a slick, flat, clean baseline, but as soon as I enter acetonitrile in the system (premixed, isocratic!) the baseline starts jumping like crazy! I cannot distinghuish concentrations of 0.5 mg/ml from the baseline. This is after leaving the system flush for > 2h and flushign the purge cell of the RID > 30 minutes.
I know that the difference in viscosity and refractive index can give problems, but I thought that if the solvent is properly mixed and degassed that should not happen anymore.
If that helps I can post some spectra with and without injection for the Zorbax (since I use the highest acetonitrile concentration there).
Hopefully I am just inexperienced and this matter can be solved very easily, I'm looking forward to your comments!
Best regards,
Adriaan