I have the most baffling problem, I am tearing out what little hair I have left. I am analyzing pesticides by 525.2, on a Thermo ISQ. I get very nice chromatography, good peak shapes, great linearity, excellent repeatability, and good performance from all of my compounds - except for Chrysene-d12, which shows random drops in response of 40 to 60 percent. It doesn't matter if the sample is an extracted drinking water or a standard freshly made from a new stock standard. Here is the oddest part - re-injection of the same solution from the same vial under the same conditions, sample or standard, will usually give excellent results, with the Chrysene-d12 coming in right where it should be. The fact that this compound is one of my internal standards makes this an even more pressing problem... totally bum-fuzzled here - any ideas?
Thermo Trace 1310, ISQ, 1uL autoinjection
30M x .25mm column in great shape
injector temp 305 (it was 285...raising it didn't seem to make a difference)
init temp 34 C, hold for .5 minutes
16 C/min to 300, hold for 3.5 minutes
MS transfer line 305 C, MS is running great, everything looks terrific - except this one compound!