Problem with left Peltier cooling on G1316A

Posted: Mon Nov 24, 2014 8:13 am
by Andreas
Dear all,

I am having trouble in our 1100 column compartment with the left peltier cooling. We cleaned it up a couple of months ago and it worked fine for a couple of months but sadly it failed again....can anybody offer a replacement part (part nr. G1316-60007) used preferably....I know Agilent but it is too expensive in Switzerland or has anybody had a similar problem and has fixed it in another way?

I would be very glad for quotes and help...even with PM.

Cheers and thanks
Andreas

Re: Problem with left Peltier cooling on G1316A

Posted: Mon Nov 24, 2014 8:42 am
by kubowicz.tomasz
Hello

90% of problems with TCC is dust inside:

-check if air outlets (on both sides) are not cover with dust - you need to remove cover
-check if both fans inside operate normally (you can replace them - they are really cheap compared to peltier elements)

Regards