Hi,

I am wondering if someone could enlighten me on if there would be a difference in results in using the

1079 Programmable Temperature Vaporizing
(PTV)
of the 1177 Split/Splitless injector (S/SL) injection port on a Varian 450-GC (with 220MS).

I am testing Volatile Fatty Acids and considering changing as i'm adding another column to the system.




Greatly appreciate any feedback,
GC newbie,

Conall
No, it depends, yes !

If you are asking whether the same sample injected to each of the inlets will give the same peak areas the answer is no.

If you are asking whether one will work better than the other (lower LOD, better rsd or whatever) it depends on what you are analysing and how.

If you are asking whether the results will be the same if you develop and validate a method for each inlet, then the answer is yes.

Peter
Peter Apps
2 posts Page 1 of 1

Who is online

In total there are 113 users online :: 0 registered, 0 hidden and 113 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 113 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry