Should it be stated in your Procedures how often you repalce the lamp?
Thanks in advance
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
Objection: For Pharma the Bottom line is that you use the lamp as long as it's appropriate for the specific analysis. An old lamp might lead to a significant higher LOQ of your methods. Related substances, trace analysis and similar therefore should include some sort of System suitability check (e.g. signal/noise of a low-concentrated sample in the specification region) to verify the lamp energy is still high enough. For other, high-concentrated stuff I'd follow your suggestion - use it until it breaks.The software should tell you that you should change your UV lamp via intensity of lamp. But bottom line you use lamp until it breaks down.