Hi All,
I am using GC Autosystem XL with FID for analysing mineral oils.
One day, flame switched off and the machine was injecting five or six samples into the column and from that time we have trouble with one ghost peak. I had changed the syringe, septum, liner with wool, o-ring (several times), precolumn and column for new one but the peak is still there! I had cleaned the space around liner with acetone, but the peak is still there.
If I am changing (increasing) the split - the peak is smaller. So, trouble should be somewhere in the injection port.
https://drive.google.com/file/d/0B1aVOl ... sp=sharing

The machine measure around 3000 samples per year and I have to say that till now, we haven't had this trouble.
Idea where could be my trouble?

Thanks a lot,

Eva
Dirty split vent line/trap ?
2 posts Page 1 of 1

Who is online

In total there are 126 users online :: 1 registered, 0 hidden and 125 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider] and 125 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry