We have 2x: Agilent 6890 network GC, 5973 MSD and a 220V supply. Both instruments have their own dedicated power line. I was investigating surge protectors and found "AC 220v Surge Brownout Voltage Protector 3600 Watts Air Conditioner Freezer". A full UPS is out of our price range at the moment, but I was wondering if the surge protector would be of any use to protect our instruments. Data loss/analysis interruption is not that big of a concern compared to replacing a board, etc. due to a nasty power surge. I was hoping someone had experience with a good, cost-effective means of protecting equipment without the investment of a UPS.



Thank you for any insight into protecting our equipment.
NR
The surge protector can only protect you against voltage spikes from your power supply (p.a. after blackout).

Sometimes there is one integrated in a power strip.

But this cannot protect you against failures of the power supply of your instruments. When the power supply is dying it is possible, that this causes a power surge on p.a. a board. Against this a surge protector can't protect your systems.
Thank you very much. That makes sense and the instrument's power supply was not a risk we had assessed. The bigger concern is the building's power supply and surges through lightning, power outage etc. Now I have a little more information to send up the chain.



NR
Thank you very much. That makes sense and the instrument's power supply was not a risk we had assessed. The bigger concern is the building's power supply and surges through lightning, power outage etc. Now I have a little more information to send up the chain.



NR
You should look into the surge protectors the power company can install on the incoming line to the building. Those are good at arresting large spikes like from lightening without costing too much I believe. We have one built into each breaker panel in our new building so we don't have to have one at each instrument. In the past ten years we have had to have three replaced because the did their job. At least they blow out first instead of passing it on down the line. This way even our ICP/MS units are protected.
The past is there to guide us into the future, not to dwell in.
you can get the same kind of surge protection by replacing your breakers to the instruments with surge protected breakers. They're not very expensive. They only trip once; you have to replace them when they blow, but they keep spikes out of your systems.
Mark Krause
Laboratory Director
Krause Analytical
Austin, TX USA
5 posts Page 1 of 1

Who is online

In total there are 27 users online :: 0 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry