Inlet pressure fluctuations and retention time shifts
Posted: Tue Nov 11, 2014 9:05 pm
I have this problem with my Agilent 6850 (GC-FID), where the flow will go down during a run (then back up to setpoint), causing shifts in retention time. It does not happen every run just every now and then during a sequence of them, which is rather odd (also does not happen consistently; that is not at regular intervals or always with the same drop values). This is what I see in the logbook when it happens:
6850A 1 Not ready:Inlet flow 24.5 at runtime 0.00 17:03:12 10/31/14
6850A 1 Not ready:Inlet pres 11.9 psi at runtime ... 17:03:12 10/31/14
6850A 1 ... 2.66 17:03:12 10/31/14
6850A 1 Not ready:Inlet pres 12.0 psi at runtime ... 17:03:13 10/31/14
6850A 1 ... 2.81 17:03:13 10/31/14
6850A 1 Not ready:Inlet pres 13.0 psi at runtime ... 17:03:13 10/31/14
6850A 1 ... 3.78 17:03:13 10/31/14
6850A 1 Not ready:Inlet pres 13.7 psi at runtime ... 17:03:13 10/31/14
6850A 1 ... 4.48 17:03:13 10/31/14
6850A 1 Not ready:Inlet pres 14.4 psi at runtime ... 17:03:13 10/31/14
6850A 1 ... 5.48 17:03:13 10/31/14
6850A 1 Not ready:Inlet pres 16.3 psi at runtime ... 17:03:13 10/31/14
6850A 1 ... 6.55 17:03:13 10/31/14
Started doing runs without injections, same issue. Checked for leaks with electronic leak detector and could not find anything at any point in the system or gas supply connections (regulator also seems to be delivering fine). Re-installed column with new ferrules, changed inlet liner, o-ring and septa. Did runs in splitless mode to see if issue was related to split ratio or split vent, but it still happened. Didn’t see any drastic temperature changes occurring during runs. Have used specific method before without any problems; also tried with different methods (tried modifying flow/pressure, temperature program, inlet/detector temperature, constant flow vs constant pressure, you name it and chances are I tried it). I was using a Restek Rtx-R1 with Integra-guard column and replaced that with just a length of guard column, but same issue.
The method I originally was running was; Inlet at 275 degC, detector at 275 degC, split ratio of 50:1 (gas saver 20 mL/min at 2.00 min), flow of 2.0 mL/min, T-programming – [65 degC held 0.5 min], [ramp (25 degC/min) to 175 hold 0.25], [ramp (30 degC/min) to 265 hold 2.25], but as I mentioned I also tried many variants of this.
Any ideas where I should be looking next?
Thanks,
Trishia
6850A 1 Not ready:Inlet flow 24.5 at runtime 0.00 17:03:12 10/31/14
6850A 1 Not ready:Inlet pres 11.9 psi at runtime ... 17:03:12 10/31/14
6850A 1 ... 2.66 17:03:12 10/31/14
6850A 1 Not ready:Inlet pres 12.0 psi at runtime ... 17:03:13 10/31/14
6850A 1 ... 2.81 17:03:13 10/31/14
6850A 1 Not ready:Inlet pres 13.0 psi at runtime ... 17:03:13 10/31/14
6850A 1 ... 3.78 17:03:13 10/31/14
6850A 1 Not ready:Inlet pres 13.7 psi at runtime ... 17:03:13 10/31/14
6850A 1 ... 4.48 17:03:13 10/31/14
6850A 1 Not ready:Inlet pres 14.4 psi at runtime ... 17:03:13 10/31/14
6850A 1 ... 5.48 17:03:13 10/31/14
6850A 1 Not ready:Inlet pres 16.3 psi at runtime ... 17:03:13 10/31/14
6850A 1 ... 6.55 17:03:13 10/31/14
Started doing runs without injections, same issue. Checked for leaks with electronic leak detector and could not find anything at any point in the system or gas supply connections (regulator also seems to be delivering fine). Re-installed column with new ferrules, changed inlet liner, o-ring and septa. Did runs in splitless mode to see if issue was related to split ratio or split vent, but it still happened. Didn’t see any drastic temperature changes occurring during runs. Have used specific method before without any problems; also tried with different methods (tried modifying flow/pressure, temperature program, inlet/detector temperature, constant flow vs constant pressure, you name it and chances are I tried it). I was using a Restek Rtx-R1 with Integra-guard column and replaced that with just a length of guard column, but same issue.
The method I originally was running was; Inlet at 275 degC, detector at 275 degC, split ratio of 50:1 (gas saver 20 mL/min at 2.00 min), flow of 2.0 mL/min, T-programming – [65 degC held 0.5 min], [ramp (25 degC/min) to 175 hold 0.25], [ramp (30 degC/min) to 265 hold 2.25], but as I mentioned I also tried many variants of this.
Any ideas where I should be looking next?
Thanks,
Trishia