Too high source pressure: Varian 1200 Triple Quadrupole
Posted: Fri Nov 07, 2014 4:37 pm
Hello everyone,
The source pressure of my MS (1200L MS/MS coupled with Varian CP-3800 GC) is too high (~500 mTorr at 80°C and ~700 mTorr at 200°C, and it should be below 150 mTorr). It's been too high since the switch of CI ion volume back to EI. Before the switch, when I turned the methane off, the pressure went down to ~30 mTorr. Then the MS was vented, ion volume changed and since then, there has been a problem.
High source pressure is likely caused by a leak. Although no changes were done to GC, I checked the fittings. They all seem to be fine. Also all the screws, seals of the vacuum interlock were cleaned/changed and it didn't help.
What is more, when the scan is run for air, there are no peaks visible, only noise. It's the same with calibration gas (no peaks are detected even with a high voltage of the electron multiplier). The intensity of the noise is lower than usual too.
I will be very grateful for any suggestions about what else can be done, checked etc.
Thank you,
Kamila
Thank you,
Kamila