Tetrachloroethylene
Posted: Thu Nov 06, 2014 9:12 am
Hello,
I have a problem with quantification of tetrachloroethylene by GC-MS. This particular method is for dichloromethane, 1,2 dichloroethane, chloroform and tetrachloroethylene. For the other 3 I have no problem, they work perfectly, but tetrachloroethylene does not have repetability, and a linear response to concentration.
The most important thing is I DO NOT use a headspace, but a L-L extraction with pentane, so the standards injected are also liquid.
The most important parameters:
Injector: 200 Celsius grades
Pre-dwell time: 0
Post-dwell time: 0
Injection volume: 2 ul
Transfer line and source: 230 Celsius grades
Quantization ion: 166
GC program: 25 Celsius grades 6 minutes, then 15 grades per minute to 120
There is not a separation problem, the peak looks awesome, but it's height and of course, area varies a lot.
I injected a Toluene-D8 as ISTD (not the perfect match) just to see if it levels the injection defects, etc. They close retention times, but the thing is Toluene-D8 is very repetable, tetrachloroethylene is not in the same injections series. Toluen-D8 will stay constant in 6 repeated injections, tetracholoroethylene will vary A LOT.
The column is Thermo TG-5SILMS which is a 5% phenyl methylpolysiloxane, 30m with 0,25 um film. I know is not the best column for VOC separation, but again, not the separation is a issue.
Why chloroform has a perfect and repetable calibration curve and tetrachloroethylene hasn't ?
I tried this:
- varied the GC program
- varied the pre and post needle dwell time
- varied split ratio (10:1, 50:1, 100:1)
- varied injector temperature (200-250 Celsius grades)
At higher or equal with 40 Celsius grades starting temperature, they will not separate. I tried to start the temperature curve from beggining, from 25 to 120 with 15 grades per min, the retention time dropped, still a good separation, but the same problem. Peak will vary a lot.
Tried to switch to ion 164, same, they are 100% proportional.
If it reacts with something to sub-speciate in column, why the other clorurated VOCs will not ? Somehow this problem is related ONLY to tetrachloroethylene.
Any ideas ? I am in a dead end.
Thank you
Vlad
I have a problem with quantification of tetrachloroethylene by GC-MS. This particular method is for dichloromethane, 1,2 dichloroethane, chloroform and tetrachloroethylene. For the other 3 I have no problem, they work perfectly, but tetrachloroethylene does not have repetability, and a linear response to concentration.
The most important thing is I DO NOT use a headspace, but a L-L extraction with pentane, so the standards injected are also liquid.
The most important parameters:
Injector: 200 Celsius grades
Pre-dwell time: 0
Post-dwell time: 0
Injection volume: 2 ul
Transfer line and source: 230 Celsius grades
Quantization ion: 166
GC program: 25 Celsius grades 6 minutes, then 15 grades per minute to 120
There is not a separation problem, the peak looks awesome, but it's height and of course, area varies a lot.
I injected a Toluene-D8 as ISTD (not the perfect match) just to see if it levels the injection defects, etc. They close retention times, but the thing is Toluene-D8 is very repetable, tetrachloroethylene is not in the same injections series. Toluen-D8 will stay constant in 6 repeated injections, tetracholoroethylene will vary A LOT.
The column is Thermo TG-5SILMS which is a 5% phenyl methylpolysiloxane, 30m with 0,25 um film. I know is not the best column for VOC separation, but again, not the separation is a issue.
Why chloroform has a perfect and repetable calibration curve and tetrachloroethylene hasn't ?
I tried this:
- varied the GC program
- varied the pre and post needle dwell time
- varied split ratio (10:1, 50:1, 100:1)
- varied injector temperature (200-250 Celsius grades)
At higher or equal with 40 Celsius grades starting temperature, they will not separate. I tried to start the temperature curve from beggining, from 25 to 120 with 15 grades per min, the retention time dropped, still a good separation, but the same problem. Peak will vary a lot.
Tried to switch to ion 164, same, they are 100% proportional.
If it reacts with something to sub-speciate in column, why the other clorurated VOCs will not ? Somehow this problem is related ONLY to tetrachloroethylene.
Any ideas ? I am in a dead end.
Thank you
Vlad