Water everywhere
Posted: Wed Nov 05, 2014 11:41 pm
6890/5973/purge&trap.
Today I cleaned the source and changed my vacuum pump oil. I haven't done a source cleaning on this particular instrument again so I ran an air & water check beforehand to get some idea of what to expect after I put the source back in. Did that, gave it a couple hours to pump down and did an air & water check.
Before:
mass 69 abundance = 543488
%water = 6.56
%nitrogen = 5.27
%oxygen = 1.28
%CO2 = 2.82
After:
mass 69 abundance = 620224
%water = 120.27
%nitrogen = 8.40
%oxygen = 2.23
%CO2 = 27.12
Maybe it just needs to pump more overnight - I'll see what it looks like in the morning. But I'm not used to seeing this on my other instruments... if I have a leak, Nitrogen is huge. I've never seen water over 10%. Any thoughts?
Today I cleaned the source and changed my vacuum pump oil. I haven't done a source cleaning on this particular instrument again so I ran an air & water check beforehand to get some idea of what to expect after I put the source back in. Did that, gave it a couple hours to pump down and did an air & water check.