I have run EPA 525.3 and can calibrate for the PAH, Adipates an Phthalates down to a low calibration of 0.01ppm with a 1ul injection full scan. Can also go down to 0.25ppb on the EPA524 Volatiles with a 5ml purge.
The volatile column allows me to take the oven temperature up to 300C so I can elute the last three PAH compounds, but it does seem to have more bleed than I am used to seeing. I was using the Rtx-502.2 but the upper limit of 270C just would not allow elution of the last PAH compounds with reasonable peak shapes or retention times.
Has anyone else used the Rxi-624SilMS column and have you noticed higher bleed on MS applications?
I also seem to have extremely high bleed from this column if I use Hydrogen carrier in it.
I was playing with the concepts in the paper from Agilent: http://hpst.cz/sites/default/files/uplo ... alyzer.pdf
My idea was to use the second column as the source for Hydrogen instead of extra flow controllers. It works really well when you do the volatile analysis and bleed in about 0.2ml/min H2 through the SVOA column. I actually found it stabilized the responses for the 1,4-Dichlorobenzene and other late eluters so that they do not increase with increasing standard concentration, which has been a problem for me for ages. But when I try it with the SVOA I get a ton of bleed from the VOA column, if I put He through the VOA column it goes away. This didn't happen when I was using the Rtx-502.2 column in the VOA side.
I am still doing some tweaking, and have found some things that you have to watch out for when doing this, but overall the setup is working great. One side benefit is you can actually use the second column and Y connector to work as a No-Vent setup when you need to change columns. Just bump up the flow of the column you are not changing and it keeps air from being sucked into the MS so you can make a column change without venting. This is the connector I am using: http://www.restek.com/catalog/view/3355 .
One thing to watch out for, you need to keep the flow of the inactive column low, or you will begin to see waves in the baseline. I think this comes from the fact that the flow control is based on there being a vacuum at the end of the column, but with the two columns flowing into a restriction, there is not a true vacuum there and the competing pressures cause fluctuations in the actual column flows. If I use greater than 0.7 ml/min on the active and less than 0.3ml/min on the inactive column the fluctuations go away. I wish I could get the 624silms column in a 40m x 0.18 configuration as it would make setting the low flow with H2 much easier, but that is something to look into in the future.
Next test will be adding pesticides to the SVOA mix and seeing how low I can go with those, so that maybe we can replace the ECD methods for drinking waters. 525.3 allows for using SIM so hopefully that will allow me to move down even further. I only wish I could use MRM for those, that would be awesome. But that will be a future project for trying to do Quechers work.
Anyone have any comments or suggestions on ways to improve this setup? I am sure there is room for improvement and I will try anything that will make it better